Fits the flight-time-to-m/z coefficients A and B from the calibration
points stored in the file.
read_sz_qtof_calibration(
path,
drop_rep = 2L,
correct = TRUE,
polarity = read_qtof_polarity(path)
)Named numeric vector with elements A and B.
Path to 'Shimadzu' .lcd file.
Which of the three replicate sets to exclude from the
fit. Defaults to 2, the factory default set described above.
Ignored when the cached correction is used, since that identifies a single
set on its own.
Logical. Whether to fold in the mass correction
'LabSolutions' applied to the run, where the file carries one. Defaults to
TRUE; the result then reproduces the vendor's mass axis, and carries a
corrected attribute to say so. FALSE returns the tuning calibration
alone, which is a few ppm out but is what the correction is expressed
against.
Ion polarity, either positive or negative, selecting
which of the two top-level entries to fit. Defaults to the polarity
recorded in the file.
The TOF Calibration Table stream holds the calibration points themselves:
pairs of a known m/z and the flight time at which the instrument actually
observed it. It is encoded as a protocol buffer, a binary format in which
every value is tagged with a number rather than a name, so the fields are
referred to below by those numbers. The top-level ones are:
10 -- positive mode (Na(NaI)n calibrants)
20 -- negative mode (I(NaI)n calibrants)
Both polarities are written to every file, so the table is the same in a positive and a negative run on the same instrument.
Within each entry are the theoretical m/z scaled by 1e9 (.10), the measured
flight time in the same units as the Centroid Data stream (.20) and a
label (.30). Five calibrants are stored per polarity, in three replicate
sets. The second is a factory default rather than a measurement --- its
flight times are byte-identical across different instruments --- and
including it inflates the error by two orders of magnitude.
'LabSolutions' fits one of the remaining sets rather than their average,
and which one varies between files. The mass correction the file caches for
the run (see read_sz_qtof_mass_correction) says which, because its offset is
tied to that set's intercept by
$$offset = B(1 - scale) + shift \times A$$
Solving that for B reproduces one set's own intercept to ~1e-7 and no
other's to better than 1e-4. That set is fitted and the correction folded
in, as \(A/scale\) and \(B - shift \times A/scale\), which reproduces
the m/z 'LabSolutions' reports to the four decimal places it writes them
with, leaving 0.03 to 0.1 ppm (median) over three files from two
instruments.
Failing that --- no cached correction, or an offset that fits no set --- the
sets named by drop_rep are dropped and the rest fitted together, which
lands within about 5 ppm. read_sz_qtof then refines that fit against the
reference ions it finds in the data.