Defines a calibration
object for the calculation of concentrations
from measurement signals including estimatations for the limit of detection
(LOD) and limit of quantification (LOQ) in accordance with DIN
32645:2008-11.
The LOD is defined as the lowest quantity of a substance that can be distinguished from the absence of that substance (blank value) within a given confidence level (alpha). The LOQ is defined as the lowest quantity of a substance that can be quantified/distinguished from another sample given with respect to a defined confidence level (k).
calibration(formula, data = NULL, model = "lm", ...)# S3 method for calibration
print(x, ...)
# S3 method for calibration
plot(x, interval = NULL, level = 0.95, ...)
lod(x, alpha = 0.01, level = 0.05)
loq(x, alpha = 0.01, k = 3, level = 0.05, maxiter = 10)
model formula providing the recorded signal intensities with
respect to the nominal analyte concentrations in the form of
signal ~ concentration
or signal ~ concentraion - 1
; model
formulae are currently restricted to those forms, however, the possibility
to use log
or sqrt
transformed data will be implemented in the
future
an optional data frame containing the variables in the model
further arguments passed to the sub method, i.e. the
respective model environment (e.g. lm
), plot
, or print
an object of class calibration
with a model formula
as shown above
Type of interval calculation (can be abbreviated); see
predict
for details
error tolerance for the detection limit (critical value)
relative uncertainty for the limit of quantification (1/beta)
a positive integer specifying the maximum number of iterations to calculate the LOQ
calibration
returns an object of class
"calibration". print
calls the function parameters together with the
respective LOD and LOQ. plot
plots the respective calibration curve
together with the measurement values. summary
may be used to retrieve
the model parameters to be found as a list item called "model".
If the data
supplied to calibration
contain more than one blank
value, i.e. measurements with a nominal concentration of zero, the LOD and
LOQ are calculated from the deviation of the blank samples. This method is
called "blank method" according to DIN 32645:2008-11 and supposed to be more
accurate than the so-called "calibration method" which will be used for the
estimation of LOD and LOQ when data
does not contain zero
concentration measurements.
DIN 32645:2008-11, 2008. Chemical analysis - Decision limit, detection limit and determination limit under repeatability conditions - Terms, methods, evaluation (Technical standard). Deutsches Institut f<U+00FC>r Normung, Berlin.
Currie, L.A., 1999. Nomenclature in evaluation of analytical methods including detection and quantification capabilities: (IUPAC Recommendations 1995). Analytica Chimica Acta 391, 105<U+2013>126.
Massart, D.L., Vandeginste, B.G., Buydens, L.M.C., Lewi, P.J., Smeyers-Verbeke, J., 1997. Handbook of chemometrics and qualimetrics: Part A. Elsevier Science Inc.
# NOT RUN {
data(din32645)
din <- calibration(Area ~ Conc, data = din32645)
din
plot(din, interval = "confidence")
summary(din$model)
lod(din)
loq(din)
# }
Run the code above in your browser using DataLab